Need help proving my HPLC doesn’t have enough precision to consistently run the same sample within 2% of itself at values of 1-5ppm
Hi lab rats, i need help proving/explaining to a group of non-scientific corporate people that my HPLC machine simply can’t consistently measure the same sample within 2% of itself.
The 2% parameter was arbitrarily set some years ago by a manager who no longer works here, and in my opinion it was not a realistic benchmark. Now it’s my task to prove in plain and clear terms why it’s not. I’m planning on just doing a simple experiment to show this, but some math/literature evidence would help too.
Im using a 1100 series Agilent HPLC and measuring levels of a compound in citrus that is typically around 2-5 ppm. We report the data to the 3rd significant digit.
I want to point out that I’m the only HPLC technician in the company, and i’ve already found MANY things in my 1 year here that we’ve had to change, so i wouldn’t be surprised if we needed to change the significant digits in our data, etc.
If you all have some suggestions, i would love to hear them! Thanks.
Posted Jul 9, 2020, 3:44 PM UTC
Hey, some quick questions about your systems and methods because I run a handful of LC/MS systems and my systems are extremely consistent but I'm not clearly understanding what the 2% you mentioned is looking at.
What sort of work are you doing, what sort of column and mp? What signals are you detecting? (DAD, ELSD, MS)?
Full disclosure, I'm not working with 1100s. I'm mostly on 1200s and infinity IIs in the Agilent department.
Feel free to DM me directly.
Hi!
So i’m doing a simple SPE extraction with ACN and Water, then running the sample, which is just citrus juice. column is a C18 reverse phase, MP is 40 ACN/60 Water, signal is VWD at one wavelength
We cross test occasionally with other labs, but they send me samples that are around 1 ppm, and 2% variance at 1 ppm is hard to hit consistently.