SpaceOAR-VUE HU override for treatment planning (y/n)?

Technical Question

Does your clinic override SpaceOAR-VUE (iodinated version of SpaceOAR) to water CT number for treatment planning as recommended by the vendor? Why or why not?

Posted Oct 19, 2021, 12:43 AM UTC

53

I would override for MV treatments. Iodine is added for visibility at kV energies, and will not have the same effect at MV energies. We do the same for rectal balloons with contrast.

6
fenpark15Therapy Physicist, PhD, DABR•

Exactly this. kV absorption is photoelectric (proportional to Z^3) where Compton MV interaction is proportional to relative electron density (and not Z). The contrast agents work for kV by increasing the effective Z of the solution, but do not have proportionally increased relative electron density.

2
IceBreaker369•

I suppose it depends on how your CT creates it's HU, and how the TPS converts from HU to tissue / density. So there could be some variation depending on how your equipped? There's no need to correct for its presence directly on the treatment, but the override will ensure the TPS doesn't try to push more dose into an area it sees as more dense.

1